IO2 nextSeg seems too weak to use for external segment sync; shutter works as frame sync

Rodolfo

New member
Hi all,
I am trying to get a reliable hardware sync signal out of a Chronos 1.4 and I am not sure whether I am misunderstanding what nextSeg is meant to do, or whether there is a firmware/hardware issue on our unit.
Camera details:
Code:
Chronos 1.4
Serial: 02236
Firmware: 0.7.0
Mode: segmented recording with external trigger
The setup is:
Code:
IO2 / AUX pin 5  -> NI USB-6363 PFI0 and/or ai7
AUX pin 6        -> DAQ common / ground
IO2 mapping      -> {source: <source>, drive: 2, debounce: false, invert: false}
The goal is to get a hardware signal from the camera that lets the DAQ confirm that each external trigger actually advanced the camera to the next segment. In other words, for a 3-segment capture, I would like to see 3 external trigger pulses into the camera and 3 corresponding camera-side acknowledgements out of IO2.

One thing that confused us at first was the drive setting. On our setup, drive=0 leaves IO2 effectively high-Z, and the DAQ reads about 0 V even when the source is set. With drive=2, static output works correctly.
Here is what we see with drive=2:
IO2 sourceCamera stateWhat we see
alwaysHighlive view~4.21 V steady
recordingsegmented mode, armed~4.21 V steady
shuttersegmented mode, armed~4.75 V pulses at about 10 kHz
shutterlive view only~0 V
nextSeg3-segment external-trigger capturePFI0 counts 0 edges; ai7 only shows small peaks around trigger times

The DAQ path itself seems fine. As a sanity check, we toggled IO2 at 1 Hz using alwaysHigh and inversion. The analogue input and the PFI0 counter both counted exactly the expected number of edges on the same wiring.
The problem appears to be specific to nextSeg.

In a 3-segment capture, the camera does fill all 3 segments and the DAQ sees the 3 external trigger pulses going into the camera. However, IO2 with source=nextSeg does not produce a usable external pulse. On the DAQ, PFI0 counts 0 edges. On ai7, the peaks near the trigger times are only about 1.5 V, which is below the USB-6363 PFI high threshold. In a separate 1 MS/s capture, we did not see any clear nextSeg pulse above about 0.5 V.

By contrast, shutter looks electrically healthy. With source=shutter, IO2 produces full-amplitude pulses while segmented recording is armed. That looks suitable as a frame-sync output, but it is not the same thing as a per-segment acknowledgement. At our current settings it pulses once per frame, not once per segment.
So I think there are two separate questions here:
  1. Is nextSeg supposed to be an externally usable TTL pulse on IO2 for each segment advance?
  2. If so, should it be full swing, similar to alwaysHigh, recording, or shutter?
  3. If nextSeg is not intended as an external signal, is there another IO2 source that gives one externally countable pulse per accepted segment trigger?
  4. Separately, is shutter / Frame Sync expected to remain valid as a full-swing frame clock during segmented armed recording?
What we would ideally like is a per-segment output pulse that is externally countable by the DAQ, for example at least 2.5 V and at least 10 µs wide. If that is not available, it would be useful to know so that we can stop chasing nextSeg and treat frame sync and segment verification as separate problems.
At the moment, our fallback is to count the trigger pulses going into the camera and then verify the camera segment count in software after capture. That works, but a hardware acknowledgement from the camera would be much cleaner.
Thanks.
RIH
 
This is an advanced use case with several questions which I recommend escalating to [email protected]

Generally speaking, although the Chronos 1.4 web interface / REST API exposes additional IO mappings such as nextSeg, the IO related settings found on the camera's LCD interface are the most reliable & supported.

It would help to see the complete ioMapping configuration you have, however I suspect you may need to leverage either io1 or io2 for the nextSeg output, and io3 for input-only.
 